Editorial Team - EMC Directory
IEC 60601-1-2 is an International EMC standard that specifies Electromagnetic emissions requirements, and Electromagnetic Immunity testing requirements & immunity test levels for Medical Electrical Equipment (ME Equipment) and Medical Electrical Systems (ME Systems). This standard is a collateral standard to the IEC 60601-1. The scope/objective of the IEC 60601-1-2 collateral standard is to outline general requirements and tests for the basic safety and essential performance of ME equipment and ME systems in the presence of electromagnetic disturbances and electromagnetic disturbances emission from ME equipment and ME systems.
Currently, the fourth edition 4.1 of this standard (IEC 60601-1-2:2014+AMD1:2020 CSV, published in 2020) is available and valid. This standard was published by the International Electrotechnical Commission (IEC) published.
This collateral standard is used to ensure that the medical equipment and medical systems will operate properly in their intended real-time electromagnetic environment without interfering with other nearby medical equipment & systems, and non-medical equipment (e.g., computers), and telecommunications (e.g., radio/TV, telephone, radio-navigation).
The medical equipment manufacturer needs to show their medical devices or products comply with the requirements of IEC 60601-1-2 to sell their products in the market.
IEC 60601-1-2 standard use and applicability:
The emission requirements and tests specified by the IEC 6061-1-2 collateral standard are generally applicable to Medical Electrical Equipment and Medical electrical systems. In this standard, tests are described based on where the ME equipment (medical device) is intended to be used. This standard requirements and tests specified are applicable for medical devices intended to be used in the following environments:
Emission measurement tests and immunity tests involved (mentioned) in IEC 60601-1-2:
EMC testing involves emission measurement testing and immunity testing. Emissions measurement testing is performed to measure the amount of EMI (electromagnetic interference or RF noise) emitted by a medical device (or medical system). The emission measurement testing ensures that the medical devices or product does not emit harmful EMI and does not cause interference with other electronic devices or systems while operating in the intended real-time environment. As per IEC 60601-1-2, the following emission measurement tests are conducted on medical devices, typically in a laboratory. During the tests, the device under test (DUT) is in its normal working (ON) state.
Immunity testing is performed in a lab environment to ensure that the medical device (or system) can withstand a specified degree of EMI from other devices while working in the intended real-time environment. In this test, test signals are applied to medical equipment/system to check the immunity of the device against EMI. The IEC 60601-1-2 collateral standard involves the following possible immunity tests as mentioned below:
The following table is provided in the IEC 60601-1-2:2014/AMD1:2020 document:
Test | Power input voltage | Power frequency |
Conducted DISTURBANCES (conducted EMISSIONS) CISPR 11 | Minimum and maximum RATED voltage | Any one frequency |
Power frequency magnetic field IMMUNITY IEC 61000-4-8 | Any one voltage | Either 50 Hz or 60 Hz. During the test, the frequency of the generated magnetic field and the power frequency of the ME equipment or ME system shall be the same |
Voltage dips IMMUNITY IEC 61000-4-11 | Minimum and maximum RATED voltage | Any one frequency |
Voltage short interruptions and voltage variations IMMUNITY IEC 61000-4-11 | Any one voltage | Any one frequency |
Proximity magnetic fields IEC 61000-4-39 | Any one voltage | Any one frequency |
What are the technical changes in the new edition of IEC 60601-1-2?
The current fourth edition 4.1 (IEC 60601-1-2:2014+AMD1:2020 CSV) cancels and replaces the third edition of IEC 60601-1-2 and features significant technical changes from the previous third edition, which are listed below.
This new edition features the following main additions:
Additional information on IEC 60601-1-2:
IEC 60601-1-2 standard requires that manufacturers of Medical devices have to submit a test plan to the accredited test laboratory prior to EMC testing. The test plan include the following details (IEC 60601-1-2; Annex G):
The document of the IEC 60601-1-2 standard contains a series of annexes that provide general guidance and include information for test plan contents, determination of immunity test levels for special environments, and immunity pass/fail criteria principles. For example, Annex E of the IEC 60601-1-2:2014+AMD1:2020 CSV provides information for the determination of immunity test levels for special environments.
Note 1: For certain types of medical electrical equipment and medical electrical systems, the requirements specified in this collateral standard might need to be modified by the special requirements of a particular standard. A particular standard is the EMC standard for a specific type of medical equipment/device (e.g., IEC 60601-2-27 is the particular standard for electrocardiographic monitoring equipment).
Note 2: As it is not practical to do so, this collateral standard does not cover all electromagnetic disturbance phenomena. Medical electrical equipment/system manufacturers need to address this during their RISK ASSESSMENT and evaluate if other electromagnetic disturbance phenomena could make their product unsafe.
Details of IEC 60601-1-2:2014+AMD1:2020 CSV document:
Publication Type | International standard |
Title | Medical electrical equipment - Part 1-2: General requirements for basic safety and essential performance - Collateral Standard: Electromagnetic disturbances - Requirements and tests |
Publication date | 2020-09-01 |
Published by | IEC (International Electrotechnical Commission) |
Edition | 4.1 |
Status | Active |
Stability date | 2028 |
Available language | English, English/French. |